quatlox said...
Lanie, I also blame the pharmaceutical companies. I would almost go to bat to say that the code is only there to insure that you are using strips that are in date. I really don't see a need to code every set of strips with a different code. I mean how does a meter made in 2005 really code a strip made in 2007. It's all "HOG WASH". I use a lot of test equipment in my lab and none of it requires codes to insure you get the same results after changing from one batch to another or even changing from one vendor to another.
Bob, You are dangerously wrong. An incorrect strip code can result in a serious measurement error. Different batches of strips do not have identical reactivities, and the code allows the meter to apply a correction factor to the result. During manufacturing, each batch of strips is sampled, and is assigned the code for the appropriate required correction factor. The error resulting from using an incorrect code depends on how much the applied correction factor differs from the right one; it can be relatively minor or very significant.
One Touch meters (InDuo, Ultra, Ultra 2, and Ultra Smart) require manual coding (1 - 49). Other meters require similar manual coding. I once coded an Ultra 2 meter to 1 and tested a strip with control solution. I repeated the test with a strip from the same batch after coding the meter to 49. One result was 7.1 and the other was 4.5.
Accu-chek Avivas do not require you to manually input the code because the code is contained on the chip that comes with each set of strips. There are other meters that also use similar techniques.
Bayer was the first company to offer self coding meters. The Bayer Ascensia Breeze and Contour don't require coding. I don't know how they do this.