Here is a
spec sheet for the ROCHE uPSA assay done on a cobas e 801.
_____________________________________
"Detection Limits:
LoB 0.006 ng/mL, LoD 0.010 ng/mL, LoQ 0.014 ng/mL (cobas e 411 analyzer, cobas e 601, cobas e 602 module)
LoB 0.006 ng/mL, LoD 0.014 ng/mL, LoQ 0.030 ng/mL (
cobas e 801 module)"
_____________________________________
For a period of time, labs using the Roche uPSA assay were told by the government they could report only the imit of quantitation (LoQ, <0.014) and not the limit of blank (LoB, <0.006); however, this has changed and the result you get if "nothing" is detected is again <0.006, as it was in the past.
For a fascinating, accessible explanation and illustration of the three terms see:
Limit of Blank, Limit of Detection and Limit of Quantitation (2008)
As I mentioned, it's best to look at the test results as they are reported by the laboratory, if available, rather than results transcribed in a non-laboratory portal.
If you had a Roche ultrasensitive psa test, the result you cited should be reported as 0.010 rather than 0.01, i.e., there should be three, not two, decimal places; this would be a transcription error. My uro's office portal once reported my uPSA result as 0.02, but changed it when I pointed out the Labcorp portal had the (correct) result, 0.020. When I first saw the 0.02, I was worried that a wrong (less sensitive) PSA test had been done.
Also keep in mind that there are
two different calibration standards for PSA testing,
WHO and
Hybritech. I'm not sure which test uses which standard, but I believe there can be a 3-17% difference between the two when reading the same sample with (non-ultrasensitive) testing. See
Comparison of Three Assays for Total and Free PSA Using Hybritech and WHO Calibrations for details. (I don't know if comparisons have been done for ultrasensitive testing, however.) This is yet another reason why you want to compare apples to apples and stick with the same test when possible.